#Teammassspec any idea what's happening every 5min in this MS2 Scan time on Astral
Comments
Log in
with your Bluesky account to leave a comment
[–]
hamishs.bsky.social
•
98 days ago
There's indeed an Astral internal calibrant step every 5 minutes whether Orbitrap Easy IC is set or not.
8
1
reply
[–]
scienthusiast.bsky.social
•
98 days ago
Thank you, @hamishs.bsky.social. That's good to know.
2
1
reply
[–]
cashwood.proteaglyco.com
•
98 days ago
Now you're only at 199.99999 Hz because of that long Astral calibration scan.
3
1
reply
[–]
scienthusiast.bsky.social
•
98 days ago
I know, the compromises we make... 😉😂
2
reply
[–]
olijhale.bsky.social
•
98 days ago
Internal calibrant ion injection? I know on Ascend we get a reagent anion AGC check every 15 mins for PTCR MSn experiments.
1
1
reply
[–]
scienthusiast.bsky.social
•
98 days ago
Easy-IC is off in the method... Also there isn't anything correlating to the spike either in TIC or MS2 spectra
0